Re: Esp + PT100

PostPosted: Sun Oct 28, 2018 6:54 am
by lucasromeiro
McChubby007 wrote:Sine you are developing this solution for a specific customer, why not price-up each option with pros/cons and give this to the customer as an R&D assessment of what they have to choose from; it is then their choice as to cost.


We do not do this for a specific client.
I have a telemetry solution company.
Some customers are looking to monitor biological freezers that reach -180C. For this temperature range I had never worked and forced us to research on the pt100 and thermocouple.
Until that happens, I had never worked with these 2. In the company we mostly use the DS18B20 sensor which has a good range for most of our applications and has a very good price!

Rudy has shown a solution to each of these and it seems that I will go this way, I will do more research.

Re: Esp + PT100

PostPosted: Sun Oct 28, 2018 3:46 pm
by McChubby007
lucasromeiro wrote:
McChubby007 wrote:Sine you are developing this solution for a specific customer, why not price-up each option with pros/cons and give this to the customer as an R&D assessment of what they have to choose from; it is then their choice as to cost.


We do not do this for a specific client.
I have a telemetry solution company.
Some customers are looking to monitor biological freezers that reach -180C. For this temperature range I had never worked and forced us to research on the pt100 and thermocouple.
Until that happens, I had never worked with these 2. In the company we mostly use the DS18B20 sensor which has a good range for most of our applications and has a very good price!

Rudy has shown a solution to each of these and it seems that I will go this way, I will do more research.

Thanks for the explanation. Being a relatively specialised requirement (such a low temperature) I agree that it seems like more research is needed, especially given the cost of special gear - sometimes you can think 'outside the box' and use components outside of what their normal application would be (example; I worked on avionics system which neeeded to measure air pressure from aircraft pitot tube, very precisely, so we used an ADC that are designed for hearts [pacemakers]).

Re: Esp + PT100

PostPosted: Mon Oct 29, 2018 12:50 pm
by lucasromeiro
McChubby007 wrote:
lucasromeiro wrote:
McChubby007 wrote:Sine you are developing this solution for a specific customer, why not price-up each option with pros/cons and give this to the customer as an R&D assessment of what they have to choose from; it is then their choice as to cost.


We do not do this for a specific client.
I have a telemetry solution company.
Some customers are looking to monitor biological freezers that reach -180C. For this temperature range I had never worked and forced us to research on the pt100 and thermocouple.
Until that happens, I had never worked with these 2. In the company we mostly use the DS18B20 sensor which has a good range for most of our applications and has a very good price!

Rudy has shown a solution to each of these and it seems that I will go this way, I will do more research.

Thanks for the explanation. Being a relatively specialised requirement (such a low temperature) I agree that it seems like more research is needed, especially given the cost of special gear - sometimes you can think 'outside the box' and use components outside of what their normal application would be (example; I worked on avionics system which neeeded to measure air pressure from aircraft pitot tube, very precisely, so we used an ADC that are designed for hearts [pacemakers]).

Nice!
Thanks for the tip!